QUESTION 5.6.1

Portable and personal gas instruments

Oil, Chemical, LPG, LNGCargo Control Room, Pumproom

What is being checked

The inspector is checking whether portable and personal gas-detection equipment is suitable for the gases and spaces concerned, within calibration and test limits, understood by users and linked to an effective response.

Ready onboard

0/6 complete

Practice questions

Answer aloud, then open the preparation point.

1Why is this portable and personal gas-detection equipment suitable?

Match the expected gases, measuring ranges and oxygen, flammable-gas and toxic-gas sensors with appropriate ranges and cross-sensitivity controls to the space and operation.

2What is the difference between a bump test and calibration?

A bump test proves sensor and alarm response to gas. Calibration adjusts accuracy against a known concentration.

3How long must a remote sample be drawn?

Use the maker's pump and hose response time plus instrument stabilisation, accounting for hose length and gas characteristics.

4What happens after an alarm or sensor fault?

Follow evacuation, ventilation, task suspension and confirmation with a suitable independent instrument; never treat a fault, inhibited channel or over-range reading as a safe atmosphere.

5Show the last complete test.

Trace inventory, calibration, bump tests, charging, issue and defect records to the identified instrument or channel and its corrective action.

Where an observation may arise

WRONG SENSOR

The instrument or oxygen, flammable-gas and toxic-gas sensors with appropriate ranges and cross-sensitivity controls did not suit the gas, range or atmosphere.

NO BUMP TEST

Users relied on a valid calibration sticker without a pre-use functional response test.

CALIBRATION GAP

Calibration gas, certificate, sensor or battery status was expired or not traceable.

SAMPLING ERROR

Hose response time, different levels, inaccessible areas or sample conditioning were not considered.

ALARM MISUNDERSTOOD

Personnel could not explain evacuation, ventilation, task suspension and confirmation with a suitable independent instrument.

SYSTEM IMPAIRED

Inspection identified failed sensor, blocked filter, leaking hose, weak battery, wrong alarm set point or insufficient instruments.

Past observations for practice

Human

The OP was not able to demonstrate the calibration process of personal gas measurement instruments.

Master's practical note

Challenge the whole measurement chain: atmosphere, sample point, hose, pump, sensor, display, alarm and user decision. A current calibration label proves only one part.

Equipment boundary and references

Equipment boundary: Use maker instructions for calibration gas, cross-sensitivity, poisoning, oxygen requirements and sampling limitations. Never use a catalytic sensor in an inert atmosphere unless approved for that purpose.

References

  • OCIMF SIRE 2.0 Question Library Part 1, Version 1.0, Question 5.6.1
  • SOLAS Chapter XI-1 regulation 7
  • OCIMF ISGOTT